Please post an mdat file with the measurements rather than screen shots. I also suggest using single sweeps. 7g looks like a lot of added mass for that driver by the way, the peak frequency only need to shift by 20% or so to get accurate results. High mass can cause odd behaviour.
Thanks. The mass is too high for REW to locate the peak in the added mass trace. Try with a much smaller added mass, in the meantime I'll see if I can make the peak search better able to cope with small peaks.
OK, thanks. I've made some modifications to the code to accommodate the low peak levels, those will be in the next beta in a week or two from now. The results with the 7g test aren't very meaningful, if you attach the 1g and 4g measurements I can run those and post the results.